A novel biomimetic artificial intervertebral disc (bioAID) replacement implant has been developed containing a swelling hydrogel representing the nucleus pulposus, a tensile strong fiber jacket as annulus fibrosus and tita-nium endplates with pins to primarily secure the device between the vertebral bodies. In this study, the design safety of this novel implant was evaluated based on several biomechanical parameters, namely compressive strength, shear-compressive strength, risk of subsidence and device expulsion as well as identifying the diurnal creep-recovery characteristics of the device. The bioAID remained intact up to 1 kN under static axial compression and only 0.4 mm of translation was observed under a compressive shear load of 20 N...