We report on tests of a radionuclide imaging system for in vivo investigations in small animals with low-energy photons as from 125-I (27–35 keV). Imaging optics features a high-resolution coded aperture mask and a fine pitch hybrid pixel detector (silicon 300- micron or 700- micron thick, or CdTe 1 mm thick) of the Medipix2 series (55 micron pitch, 256X256 pixels). The coded aperture had 480 70- micron holes in 100- micron-thick tungsten. Laboratory tests with a 109-Cd 22 keV source and a microfocus X-ray tube (35 kVp, Mo anode) show a system resolution of about 110 micron at magnification m= 2.12 and a sensitivity improvement of 30:1 as compared to a 300- micron pinhole collimator. The field of view also depends on magnification: in the e...