The NEMA NU-2 standard describes a protocol for measurement of scatter fraction (SF) using an axially-aligned line source, offset at 45 mm from the central axis, in a cylindrical polyethylene phantom. In this work, which is an extension of our preliminary results previuosly published in the Proceedings of IEEE NSS/MIC 2018 [1], we aim to evaluate the performance of the NEMA NU-2 SF protocol in a Siemens Biograph mCT PET/CT whole-body scanner and a long axial field-of-view (LAFOV) total-body PET scanner to determine whether modifications to the NEMA NU-2 SF protocol are needed for the characterisation of scatter in such scanners. In addition, we evaluate the impact of patient body mass index (BMI) on SF in a LAFOV scanner. The Siemens Biogra...