Monitoring of the helium flow into a gas chromatograph (GC) by means of an electronic flow meter has been used to optimize and control large-volume on-column injections. The nature of the observed carrier gas flow-rate profiles is discussed in some detail. A rather strong dependence of the evaporation rate on the injection speed was found for injections into a 0.32 mm I.D. retention gap, which can be attributed to the pressure drop along the retention gap when using a solvent vapour exit (SVE). The variation of the evaporation rate with the injection speed was found to be less critical for a 0.53 mm I.D. retention gap. The carrier gas flow-rate profile during the actual injection was used to effect the automated closure of the SVE precisely...