International audiencePurpose: Accurate reconstruction of the dose delivered by 90Y microspheres using a postembolization PET scan would permit the establishment of more accurate dose–response relationships for treatment of hepatocellular carcinoma with 90Y. However, the quality of the PET data obtained is compromised by several factors, including poor count statistics and a very high random fraction. This work uses Monte Carlo simulations to investigate what impact factors other than low count statistics have on the quantification of 90Y PET.Methods: PET acquisitions of two phantoms—a NEMA PET phantom and the NEMA IEC PET body phantom-containing either 90Y or 18F were simulated using GATE. Simulated projections were created with subsets of...