Efforts have been made to design electrodes that significantly reduce not only the acute and chronic stimulation thresholds, but also attenuate the early peaking phenomenon and polarization. At two voltage levels (2.7 V and 5.4 V, respectively), we evaluated the right ventricular stimulation thresholds obtained with a new, iridium oxide-coated electrode in ten patients who received a VVI pacemaker. Measurements were made at implant and at multiple intervals for 1 year. Pulse width stimulation thresholds at implant were as follow: 0.04 +/- 0.008 msec at 2.7 V, 0.03 +/- 0.004 msec at 5.4 V; values at 2 weeks were 0.14 +/- 0.06 msec at 2.7 V, 0.07 +/- 0.025 msec at 5.4 V; values at 3 months were 0.09 +/- 0.03 msec at 2.7 V, 0.05 +/- 0.01 msec ...