Contingency tables showing agreement in case counts across the STANDARD G6PD Test and the reference assay using the manufacturer’s threshold values at 30% and 70% G6PD activity thresholds, on A) capillary specimens, B) venous specimens (excluding contrived), and C) Venous specimens (including contrived).</p
* The UK study, which included newborns and samples with blood disorders, and the Thailand study, wh...
Summary diagnostic performance of the STANDARD G6PD Test for identifying G6PD deficient males and fe...
Performance indicators applying the 6.0 U/g Hb threshold on the STANDARD G6PD Test for deficient and...
3x3 agreement tables between the STANDARD G6PD Test and the reference assay G6PD percent activity, o...
Percent agreement between the STANDARD G6PD Test and the spectrophotometric reference test using the...
Contingency tables showing agreement in classification of anemia status between the STANDARD G6PD Te...
2x2 tables for the clinical sensitivity and specificity of the STANDARD G6PD Test: A) for G6PD-defic...
Percent agreement using the manufacturer’s thresholds at 30% and 80% G6PD activity for A) venous spe...
Regression analysis of STANDARD G6PD Test’s G6PD activity on A) venous specimens and B) capillary sp...
Diagnostic performance of the STANDARD G6PD Test using the manufacturer’s thresholds compared with n...
G6PD status classification measured by the STANDARD G6PD Test is shown on the X-axis, for capillary ...
30% and 70% normal G6PD activity thresholds were used to classify G6PD deficient and intermediate st...
Regression analysis of STANDARD G6PD Test total hemoglobin measurement on A) capillary specimens, an...
Regression analyses and Bland-Altman plots of STANDARD G6PD Test’s G6PD activity on A) venous specim...
The frequency is plotted in 10% increments for males and females. The 30% and 70% activity limits ar...
* The UK study, which included newborns and samples with blood disorders, and the Thailand study, wh...
Summary diagnostic performance of the STANDARD G6PD Test for identifying G6PD deficient males and fe...
Performance indicators applying the 6.0 U/g Hb threshold on the STANDARD G6PD Test for deficient and...
3x3 agreement tables between the STANDARD G6PD Test and the reference assay G6PD percent activity, o...
Percent agreement between the STANDARD G6PD Test and the spectrophotometric reference test using the...
Contingency tables showing agreement in classification of anemia status between the STANDARD G6PD Te...
2x2 tables for the clinical sensitivity and specificity of the STANDARD G6PD Test: A) for G6PD-defic...
Percent agreement using the manufacturer’s thresholds at 30% and 80% G6PD activity for A) venous spe...
Regression analysis of STANDARD G6PD Test’s G6PD activity on A) venous specimens and B) capillary sp...
Diagnostic performance of the STANDARD G6PD Test using the manufacturer’s thresholds compared with n...
G6PD status classification measured by the STANDARD G6PD Test is shown on the X-axis, for capillary ...
30% and 70% normal G6PD activity thresholds were used to classify G6PD deficient and intermediate st...
Regression analysis of STANDARD G6PD Test total hemoglobin measurement on A) capillary specimens, an...
Regression analyses and Bland-Altman plots of STANDARD G6PD Test’s G6PD activity on A) venous specim...
The frequency is plotted in 10% increments for males and females. The 30% and 70% activity limits ar...
* The UK study, which included newborns and samples with blood disorders, and the Thailand study, wh...
Summary diagnostic performance of the STANDARD G6PD Test for identifying G6PD deficient males and fe...
Performance indicators applying the 6.0 U/g Hb threshold on the STANDARD G6PD Test for deficient and...