Introduction: The purpose of this study was to evaluate the effect of autoclave sterilization on cyclic fatigue resistance of rotary endodontic instruments made of traditional and new nickel-titanium (NiTi) alloys. Methods: Four NiTi rotary endodontic instruments of the same size (tip diameter 0.40 mm and constant .04 taper) were selected: K3, Mtwo, Vortex, and K3 XF prototypes. Each group was then divided into 2 subgroups, unsterilized instruments and sterilized instruments. The sterilized instruments were subjected to 10 cycles of autoclave sterilization. Twelve files from each different subgroup were tested for cyclic fatigue resistance. Means and standard deviations of number of cycles to failure (NCF) and fragment length of the fractur...