Constant-amplitude uniaxial tension-compression fatigue tests were conducted on specimens fabricated from 12 sets of acrylic bone cements, covering cement formulations with three different viscosities (so-called high- , medium- and low-viscosity varieties), two different methods of mixing the cement constituents (so-called hand- and vacuum-mixed methods) and two test specimen shapes (rectangular-cross-sectioned or flat and circular-cross-sectioned or round ). The test results-namely, the number of fatigue stress cycles, Nf-were analyzed using the linearized transformation of the three-parameter Weibull relationship, allowing the values of the Weibull mean, NWM, to be determined for each set. Values ranged from 14,300 to 1,284,33...
Different cyclic loading modes have been used in in vitro fatigue studies of PMMA bone cement. It is...
The thrust of the present work was the experimental determination of the uniaxial static compressive...
Fatigue life of Simplex P bone cement was tested at three different stress amplitudes by using speci...
Constant-amplitude uniaxial tension-compression fatigue tests were conducted on specimens fabricated...
The goal of the present work was to investigate the influence of the viscosity classification of an ...
There is no consensus over the optimal criterion to define the fatigue life of bone cement in vitro....
The goal of the present work was to test the hypothesis that test frequency, f, does not have a stat...
The fatigue performance of bone cement is influenced by the testing parameters. In previous in vitro...
Two different methods have been used to fatigue test four bone cements. Each method has been used pr...
Over the past three decades or so, a very large volume of literature has been generated on the impac...
In the literature on fatigue testing of acrylic bone cements, data sets of various sizes have been u...
The goal of the present work was to establish the relative influence of one exogenous variable versu...
Bone cements are a common tool for the fixation of high load bearing endoprostheses. One of the most...
The weight-average molecular weight (MW(w)) of a cement and the method used to mix its powder and li...
Two different methods have been used to fatigue test four bone cements. Each method has been used pr...
Different cyclic loading modes have been used in in vitro fatigue studies of PMMA bone cement. It is...
The thrust of the present work was the experimental determination of the uniaxial static compressive...
Fatigue life of Simplex P bone cement was tested at three different stress amplitudes by using speci...
Constant-amplitude uniaxial tension-compression fatigue tests were conducted on specimens fabricated...
The goal of the present work was to investigate the influence of the viscosity classification of an ...
There is no consensus over the optimal criterion to define the fatigue life of bone cement in vitro....
The goal of the present work was to test the hypothesis that test frequency, f, does not have a stat...
The fatigue performance of bone cement is influenced by the testing parameters. In previous in vitro...
Two different methods have been used to fatigue test four bone cements. Each method has been used pr...
Over the past three decades or so, a very large volume of literature has been generated on the impac...
In the literature on fatigue testing of acrylic bone cements, data sets of various sizes have been u...
The goal of the present work was to establish the relative influence of one exogenous variable versu...
Bone cements are a common tool for the fixation of high load bearing endoprostheses. One of the most...
The weight-average molecular weight (MW(w)) of a cement and the method used to mix its powder and li...
Two different methods have been used to fatigue test four bone cements. Each method has been used pr...
Different cyclic loading modes have been used in in vitro fatigue studies of PMMA bone cement. It is...
The thrust of the present work was the experimental determination of the uniaxial static compressive...
Fatigue life of Simplex P bone cement was tested at three different stress amplitudes by using speci...