Magnetic resonance-guided radiotherapy technology is relatively new and commissioning publications, quality assurance (QA) protocols and commercial products are limited. This work provides guidance for implementation measurements that may be performed on the Elekta Unity MR-Linac (Elekta, Stockholm, Sweden). Adaptations of vendor supplied phantoms facilitated determination of gantry angle accuracy and linac isocentre, whereas in-house developed phantoms were used for end-to-end testing and anterior coil attenuation measurements. Third-party devices were used for measuring beam quality, reference dosimetry and during treatment plan commissioning; however, due to several challenges, variations on standard techniques were required. Gantry angl...
The introduction of real-time imaging by magnetic resonance guided linear accelerators (MR-Linacs) e...
Marcus Powers commissioned the Southern Hemisphere's first clinical MRI-guided Linear Accelerator an...
Methods and results for commissioning of the complete VMAT delivery chain are presented for the comb...
Magnetic resonance-guided radiotherapy technology is relatively new and commissioning publications, ...
BACKGROUND:Magnetic Resonance linear accelerator (MR-linac) systems represent a new type of technolo...
© Copyright © 2020 Jelen, Dong, Begg, Roberts, Whelan, Keall and Liney. Purpose: Unique characterist...
Purpose: A recently installed MR-linac couples real-time, non-ionizing tumor tracking with double st...
Purpose: The purpose of this paper is to describe a practical approach to commissioning and quality ...
In radiotherapy the main goal is to kill cancer cells, prevent them from growing and reproducing, wh...
Dosimetric quality assurance (QA) of the new Elekta Unity (MR-linac) will differ from the QA perform...
Radiotherapy treatments need adequate quality control (QC) to ensure a correct delivery of the presc...
Purpose: Recently, MR-linacs have been integrated into clinical practice, introducing new needs for ...
To perform patient plan-quality assurance (QA) on the newly installed MR-Linac (MRL) there was a nee...
Magnetic resonance (MR)-guided radiotherapy involves complex irradiation- and imaging devices (MR-Li...
Consistent quality assurance (QA) programs are vital to MR-guided radiotherapy (MRgRT), for ensuring...
The introduction of real-time imaging by magnetic resonance guided linear accelerators (MR-Linacs) e...
Marcus Powers commissioned the Southern Hemisphere's first clinical MRI-guided Linear Accelerator an...
Methods and results for commissioning of the complete VMAT delivery chain are presented for the comb...
Magnetic resonance-guided radiotherapy technology is relatively new and commissioning publications, ...
BACKGROUND:Magnetic Resonance linear accelerator (MR-linac) systems represent a new type of technolo...
© Copyright © 2020 Jelen, Dong, Begg, Roberts, Whelan, Keall and Liney. Purpose: Unique characterist...
Purpose: A recently installed MR-linac couples real-time, non-ionizing tumor tracking with double st...
Purpose: The purpose of this paper is to describe a practical approach to commissioning and quality ...
In radiotherapy the main goal is to kill cancer cells, prevent them from growing and reproducing, wh...
Dosimetric quality assurance (QA) of the new Elekta Unity (MR-linac) will differ from the QA perform...
Radiotherapy treatments need adequate quality control (QC) to ensure a correct delivery of the presc...
Purpose: Recently, MR-linacs have been integrated into clinical practice, introducing new needs for ...
To perform patient plan-quality assurance (QA) on the newly installed MR-Linac (MRL) there was a nee...
Magnetic resonance (MR)-guided radiotherapy involves complex irradiation- and imaging devices (MR-Li...
Consistent quality assurance (QA) programs are vital to MR-guided radiotherapy (MRgRT), for ensuring...
The introduction of real-time imaging by magnetic resonance guided linear accelerators (MR-Linacs) e...
Marcus Powers commissioned the Southern Hemisphere's first clinical MRI-guided Linear Accelerator an...
Methods and results for commissioning of the complete VMAT delivery chain are presented for the comb...