Temperature rise during removal of fractured components out of the implant body:an in vitro study comparing two ultrasonic devices and five implant types

  • Meisberger, Eric W.
  • Bakker, Sjoerd J. G.
  • Cune, Marco S.
Open PDF
Publication date
December 2015
Language
English

Abstract

Background: Ultrasonic instrumentation under magnification may facilitate mobilization of screw remnants but may induce heat trauma to surrounding bone. An increase of 5 degrees C is considered detrimental to osseointegration. The objective of this investigation was to examine the rise in temperature of the outer implant body after 30 s of ultrasonic instrumentation to the inner part, in relation to implant type, type of ultrasonic equipment, and the use of coolants in vitro.Methods: Two ultrasonic devices (Satelec Suprasson T Max and Electro Medical Systems (EMS) miniMaster) were used on five different implant types that were provided with a thermo couple (Astra 3.5 mm, bone level Regular CrossFit (RC) 4.1 mm, bone level Narrow CrossFit (N...