OBJECTIVES Calculation of accurate T1 relaxivity (r1) values for gadolinium-based magnetic resonance contrast agents (GBCAs) is a complex process. As such, often referenced r1 values for the GBCAs at 1.5 T, 3 T, and 7 T are based on measurements obtained in media that are not clinically relevant, derived from only a small number of concentrations, or available for only a limited number of GBCAs. This study derives the r1 values of the 8 commercially available GBCAs in human whole blood at 1.5 T, 3 T, and 7 T. MATERIALS AND METHODS Eight GBCAs were serially diluted in human whole blood, at 7 concentrations from 0.0625 to 4 mM. A custom-built phantom held the dilutions in air-tight cylindrical tubes maintained at 37 ± 0.5°C by a heat-circu...
Magnetic resonance imaging (MRI) is a non-invasive imaging modality which offers excellent spatial r...
PurposeTo demonstrate that constant coefficient of variation (CV), but nonconstant absolute variance...
T1 mapping is crucial for many quantitative magnetic resonance imaging (MRI) procedures. However, st...
Background The availability of data in the medical literature for the T relaxivities of the Gd-based...
Dynamic susceptibility contrast (DSC) MRI is clinically used to measure brain perfusion by monitorin...
PURPOSE: Accurate characterization of contrast reagent (CR) longitudinal relaxivity in whole blood i...
MRI contrast agents (CA) whose contrast enhancement remains relatively high even at the higher end o...
This study has measured the longitudinal and transverse (T2* relaxivity curves for ProHance (Gadoter...
OBJECTIVE: Our aim was to study the influence of small variations in spatial resolution and contrast...
<div><p>Abstract Objective: To compare an albumin-bound gadolinium chelate (gadofosveset trisodium)...
The question of improved relaxivity, and potential efficacy therein, for a next-generation of magnet...
contrast agents were introduced into the clinical environment, and multiple in vivo studies compared...
Most contrast agents for magnetic respnance imaging (MRI) are gadolinium-based T1 shortening agents....
The relaxivity of commercially available gadolinium (Gd)-based contrast agents was studied for X-nuc...
<p>Measurements were performed at 1.5 Tesla MR field strength in bovine plasma at 37°C. For each lip...
Magnetic resonance imaging (MRI) is a non-invasive imaging modality which offers excellent spatial r...
PurposeTo demonstrate that constant coefficient of variation (CV), but nonconstant absolute variance...
T1 mapping is crucial for many quantitative magnetic resonance imaging (MRI) procedures. However, st...
Background The availability of data in the medical literature for the T relaxivities of the Gd-based...
Dynamic susceptibility contrast (DSC) MRI is clinically used to measure brain perfusion by monitorin...
PURPOSE: Accurate characterization of contrast reagent (CR) longitudinal relaxivity in whole blood i...
MRI contrast agents (CA) whose contrast enhancement remains relatively high even at the higher end o...
This study has measured the longitudinal and transverse (T2* relaxivity curves for ProHance (Gadoter...
OBJECTIVE: Our aim was to study the influence of small variations in spatial resolution and contrast...
<div><p>Abstract Objective: To compare an albumin-bound gadolinium chelate (gadofosveset trisodium)...
The question of improved relaxivity, and potential efficacy therein, for a next-generation of magnet...
contrast agents were introduced into the clinical environment, and multiple in vivo studies compared...
Most contrast agents for magnetic respnance imaging (MRI) are gadolinium-based T1 shortening agents....
The relaxivity of commercially available gadolinium (Gd)-based contrast agents was studied for X-nuc...
<p>Measurements were performed at 1.5 Tesla MR field strength in bovine plasma at 37°C. For each lip...