This study aimed to describe the commissioning of small field size radiosurgery cones in a 6-MV flattening filter free (FFF) beam and report our measured values. Four radiosurgery cones of diameters 5, 10, 12.5, and 15¿mm supplied by Elekta Medical were commissioned in a 6-MV FFF beam from an Elekta Versa linear accelerator. The extraction of a reference signal for measuring small fields in scanning mode is challenging. A transmission chamber was attached to the lower part of the collimators and used for percentage depth dose (PDD) and profile measurements in scanning mode with a stereotactic diode. Tissue-maximum ratios (TMR) and output factors (OF) for all collimators were measured with a stereotactic diode (IBA). TMR and the OF for the l...
Physical aspects of very small diameter X-ray beams used for a linac-based stereotactic radiosurgery...
<div><p>Silicon-diode-based detectors are commonly used for the dosimetry of small radiotherapy beam...
This project presents recent advances in small field dosimetry for radiosurgery treatments with Brai...
Purpose: To evaluate the response of the four smallest active volume thimble type ionization chamber...
International audiencePurpose The use of small photon fields is now an established practice in stere...
Purpose: Conical collimators, or cones, are tertiary collimators that attach to a radiotherapy linac...
Introduction: Using the small field in modern radiotherapy, the present study aimed at measuring the...
Purpose Small field x-ray beam dosimetry is difficult due to a lack of lateral electronic equilibriu...
Purpose: This work introduces the concept of very small field size. Output factor (OPF) measurements...
Les nouvelles techniques de la radiothérapie (Stéréotaxie, IMRT, VMAT, IGRT...etc) utilisent des fai...
Purpose: Flattening filter free (FFF) beams are increasingly being considered for stereotactic radio...
846-850This investigation aimed to measure and commission small unflat photon beams using the detect...
Background: Beam matching is widely used to ensure that linear accelerators used in radiotherapy hav...
PURPOSE: Stereotactic radiosurgery (SRS) can be delivered with a standard linear accelerator (linac)...
Purpose: We evaluated the energy responses of a radiophotoluminescent glass dosimeter (RPLD) to vari...
Physical aspects of very small diameter X-ray beams used for a linac-based stereotactic radiosurgery...
<div><p>Silicon-diode-based detectors are commonly used for the dosimetry of small radiotherapy beam...
This project presents recent advances in small field dosimetry for radiosurgery treatments with Brai...
Purpose: To evaluate the response of the four smallest active volume thimble type ionization chamber...
International audiencePurpose The use of small photon fields is now an established practice in stere...
Purpose: Conical collimators, or cones, are tertiary collimators that attach to a radiotherapy linac...
Introduction: Using the small field in modern radiotherapy, the present study aimed at measuring the...
Purpose Small field x-ray beam dosimetry is difficult due to a lack of lateral electronic equilibriu...
Purpose: This work introduces the concept of very small field size. Output factor (OPF) measurements...
Les nouvelles techniques de la radiothérapie (Stéréotaxie, IMRT, VMAT, IGRT...etc) utilisent des fai...
Purpose: Flattening filter free (FFF) beams are increasingly being considered for stereotactic radio...
846-850This investigation aimed to measure and commission small unflat photon beams using the detect...
Background: Beam matching is widely used to ensure that linear accelerators used in radiotherapy hav...
PURPOSE: Stereotactic radiosurgery (SRS) can be delivered with a standard linear accelerator (linac)...
Purpose: We evaluated the energy responses of a radiophotoluminescent glass dosimeter (RPLD) to vari...
Physical aspects of very small diameter X-ray beams used for a linac-based stereotactic radiosurgery...
<div><p>Silicon-diode-based detectors are commonly used for the dosimetry of small radiotherapy beam...
This project presents recent advances in small field dosimetry for radiosurgery treatments with Brai...