BACKGROUND: To describe the histological features of failed gold micro shunts (GMS) in unsuccessful implantations for refractory primary open-angle glaucoma (POAG). METHODS: This was an interventional case series study. Five eyes of five glaucomatous patients with unsuccessful GMS implantation underwent shunt removal. Each device was sectioned into three portions: proximal or anterior chamber (AC) portion, middle or scleral portion and distal or suprachoroidal (SC) portion. The histological analysis was performed throughout the whole extent of the shunt, describing both the inner spaces and the outer surface. RESULTS: At the moment of removal all devices were correctly located into the SC space and in AC, with the exception of a case p...
Aim: The aim of this report is to explore the risk factors of XEN stent obstruction, suggesting the ...
Purpose Ahmed glaucoma valve (AGV) implant is an aqueous shunt device used to control intraocu-lar p...
We report a case of an explanted stainless steel miniature glaucoma drainage device (EX-PRESS) impla...
BACKGROUND: To describe the histological features of failed gold micro shunts (GMS) in unsuccessful ...
AIMS:To describe the conjunctival epithelial features seen with in vivo confocal microscopy (IVCM) a...
morphological evaluation of gold micro shunt after unsuccessful surgical treatment of patients with ...
AIMS: This study aimed to evaluate the efficacy of Gold Micro Shunt (GMS) for suprachoroidal drainag...
Purpose: To describe an unusual case of glaucoma drainage tube blockage. Patients: A 76-year-old wom...
Ryan Le,1 Neeru Gupta1–3 1Keenan Research Center, Li Ka Shing Knowledge Institute, St. Michael...
PURPOSE To describe an unusual case of glaucoma drainage tube blockage. PATIENTS A 76-year-...
To evaluate the 10-year follow-up efficacy and safety of gold micro shunt (GMS) in patients with ref...
Purpose: The aim of this study was to assess the effect of various tube parameters on corneal endoth...
PURPOSE:Ahmed glaucoma valve (AGV) implant is an aqueous shunt device used to control intraocular pr...
Introduction: The iStent inject (R) (Glaukos Corporation, CA, USA) is a relatively new device design...
Glaucoma is among the most prevalent causes of blindness, resulting from increase in fluid pressure ...
Aim: The aim of this report is to explore the risk factors of XEN stent obstruction, suggesting the ...
Purpose Ahmed glaucoma valve (AGV) implant is an aqueous shunt device used to control intraocu-lar p...
We report a case of an explanted stainless steel miniature glaucoma drainage device (EX-PRESS) impla...
BACKGROUND: To describe the histological features of failed gold micro shunts (GMS) in unsuccessful ...
AIMS:To describe the conjunctival epithelial features seen with in vivo confocal microscopy (IVCM) a...
morphological evaluation of gold micro shunt after unsuccessful surgical treatment of patients with ...
AIMS: This study aimed to evaluate the efficacy of Gold Micro Shunt (GMS) for suprachoroidal drainag...
Purpose: To describe an unusual case of glaucoma drainage tube blockage. Patients: A 76-year-old wom...
Ryan Le,1 Neeru Gupta1–3 1Keenan Research Center, Li Ka Shing Knowledge Institute, St. Michael...
PURPOSE To describe an unusual case of glaucoma drainage tube blockage. PATIENTS A 76-year-...
To evaluate the 10-year follow-up efficacy and safety of gold micro shunt (GMS) in patients with ref...
Purpose: The aim of this study was to assess the effect of various tube parameters on corneal endoth...
PURPOSE:Ahmed glaucoma valve (AGV) implant is an aqueous shunt device used to control intraocular pr...
Introduction: The iStent inject (R) (Glaukos Corporation, CA, USA) is a relatively new device design...
Glaucoma is among the most prevalent causes of blindness, resulting from increase in fluid pressure ...
Aim: The aim of this report is to explore the risk factors of XEN stent obstruction, suggesting the ...
Purpose Ahmed glaucoma valve (AGV) implant is an aqueous shunt device used to control intraocu-lar p...
We report a case of an explanted stainless steel miniature glaucoma drainage device (EX-PRESS) impla...