International audienceUltrasonic backscatter coefficient (BSC) measurements were performed on K562 cell pellet biophantoms with cell concentrations ranging from 0.006 to 0.30 in the 10-42 MHz frequency bandwidth. Three scattering models, namely the Fluid-Filled Sphere Model (FFSM), the Particle Model (PM) and the Structure Factor Model (SFM), were compared for modeling the scattering from an ensemble of concentrated cells. A parameter estimation procedure was developed in order to estimate the scatterer size and relative impedance contrast that could explain the measured BSCs from all the studied cell concentrations. This procedure was applied to the BSC data from K562 cell pellet biophantoms in the 10-42 MHz frequency bandwidth and to the ...
International audienceQuantitative ultrasound technique utilizes the amplitude and frequency depende...
In this study a cell pellet biophantom technique is introduced and applied to the ultrasonic backsca...
International audienceQuantitative ultrasound (QUS) methods based on the backscatter coefficient (BS...
International audienceUltrasonic backscatter coefficient (BSC) measurements were performed on K562 c...
International audienceUltrasonic backscatter coefficient (BSC) measurements were performed on K562 c...
International audienceUltrasonic backscatter coefficient (BSC) measurements were performed on K562 c...
International audienceUltrasonic backscatter coefficient (BSC) measurements were performed on K562 c...
International audienceUltrasonic backscatter coefficient (BSC) measurements were performed on K562 c...
International audienceUltrasonic backscatter coefficient (BSC) measurements were performed on K562 c...
International audienceUltrasonic backscatter coefficient (BSC) measurements were performed on K562 c...
International audienceQuantitative Ultrasound techniques for determining the tissue microstructure r...
International audienceQuantitative Ultrasound techniques for determining the tissue microstructure r...
International audienceUltrasonic backscatter coefficient (BSC) measurements were performed on K562 c...
International audienceUltrasonic backscatter coefficient (BSC) measurements were performed on K562 c...
International audienceQuantitative ultrasound technique utilizes the amplitude and frequency depende...
International audienceQuantitative ultrasound technique utilizes the amplitude and frequency depende...
In this study a cell pellet biophantom technique is introduced and applied to the ultrasonic backsca...
International audienceQuantitative ultrasound (QUS) methods based on the backscatter coefficient (BS...
International audienceUltrasonic backscatter coefficient (BSC) measurements were performed on K562 c...
International audienceUltrasonic backscatter coefficient (BSC) measurements were performed on K562 c...
International audienceUltrasonic backscatter coefficient (BSC) measurements were performed on K562 c...
International audienceUltrasonic backscatter coefficient (BSC) measurements were performed on K562 c...
International audienceUltrasonic backscatter coefficient (BSC) measurements were performed on K562 c...
International audienceUltrasonic backscatter coefficient (BSC) measurements were performed on K562 c...
International audienceUltrasonic backscatter coefficient (BSC) measurements were performed on K562 c...
International audienceQuantitative Ultrasound techniques for determining the tissue microstructure r...
International audienceQuantitative Ultrasound techniques for determining the tissue microstructure r...
International audienceUltrasonic backscatter coefficient (BSC) measurements were performed on K562 c...
International audienceUltrasonic backscatter coefficient (BSC) measurements were performed on K562 c...
International audienceQuantitative ultrasound technique utilizes the amplitude and frequency depende...
International audienceQuantitative ultrasound technique utilizes the amplitude and frequency depende...
In this study a cell pellet biophantom technique is introduced and applied to the ultrasonic backsca...
International audienceQuantitative ultrasound (QUS) methods based on the backscatter coefficient (BS...