Background: Different target-filter combinations in computed radiography have different impacts on the dose and image quality in digital radiography. This study aims to evaluate the mean glandular dose (MGD) and modulation transfer function (MTF) of various target-filter combinations by investigating the signal intensities of X-ray beams. Methods: General Electric (GE) Senographe DMR Plus mammography unit was used for MGD and MTF evaluation. The measured MGD was compared with the dose reference level (DRL), whereas the MTF was evaluated using ImageJ 1.46o software. A modified Mammography Accreditation Phantom RMI 156 was exposed using different target-filter combinations of molybdenum-molybdenum (Mo-Mo), molybdenum-rhodium (Mo-Rh) and rhodi...
The performance of a digital mammography system (Siemens Mammomat Novation) using different target/f...
Evaluating the patient dose or exposure parameters considering the image quality can improve the cha...
Introduction: Image quality of digital mammography system is generally defined by three primary phys...
Background: Different target-filter combinations in computed radiography have different impacts on t...
We had conducted a survey on Mean Glandular Dose (MGD) from Full-Field Digital Mammography systems ...
Recently, digital mammography units have been introduced into clinical use in Japan and patients und...
The mean glandular dose (MGD) was measured for a breast phantom by using molybdenum/molybdenum and ...
Introduction: Average glandular dose calculation in mammography with Mo-Rh target-filter and dose ca...
To develop tables of normalized glandular dose coefficients D(g)N for a range of anode-filter combin...
The aim of this study was to compare mean glandular dose (MGD) in all full-field digital mammography...
In performing breast screening, a mammography must be capable of imaging microcalcifications with th...
In digital breast tomosynthesis and digital mammography, the x-ray beam filter material and thicknes...
Mammography refers to the X-ray examination of the human breast, and is considered the single most i...
Abstract: In performing breast screening, a mammography must be capable of imaging microcalcificatio...
"Major Subject: Nuclear Engineering".Conversion factors relating entrance skin exposure to average g...
The performance of a digital mammography system (Siemens Mammomat Novation) using different target/f...
Evaluating the patient dose or exposure parameters considering the image quality can improve the cha...
Introduction: Image quality of digital mammography system is generally defined by three primary phys...
Background: Different target-filter combinations in computed radiography have different impacts on t...
We had conducted a survey on Mean Glandular Dose (MGD) from Full-Field Digital Mammography systems ...
Recently, digital mammography units have been introduced into clinical use in Japan and patients und...
The mean glandular dose (MGD) was measured for a breast phantom by using molybdenum/molybdenum and ...
Introduction: Average glandular dose calculation in mammography with Mo-Rh target-filter and dose ca...
To develop tables of normalized glandular dose coefficients D(g)N for a range of anode-filter combin...
The aim of this study was to compare mean glandular dose (MGD) in all full-field digital mammography...
In performing breast screening, a mammography must be capable of imaging microcalcifications with th...
In digital breast tomosynthesis and digital mammography, the x-ray beam filter material and thicknes...
Mammography refers to the X-ray examination of the human breast, and is considered the single most i...
Abstract: In performing breast screening, a mammography must be capable of imaging microcalcificatio...
"Major Subject: Nuclear Engineering".Conversion factors relating entrance skin exposure to average g...
The performance of a digital mammography system (Siemens Mammomat Novation) using different target/f...
Evaluating the patient dose or exposure parameters considering the image quality can improve the cha...
Introduction: Image quality of digital mammography system is generally defined by three primary phys...