Extensive cross-reference measurements of well-known frequency ratios using various sizes of carbon cluster ions 12Cn+ (10 ≤ n ≤ 23) were performed to determine the effects limiting the accuracy of mass measurements at the Penning-trap facility TRIGA-TRAP. Two major contributions to the uncertainty of a mass measurement have been identified. Fluctuations of the magnetic field cause an uncertainty in the frequency ratio due to the required calibration by a reference ion of uf(νref)/νref = 6(2) × 10-11/min × Δ t. A mass-dependent systematic shift of the frequency ratio of ϵm(r)/r = -2.2(2) × 10-9 × (m-mref)/u has been found as well. Finally, the nuclide 197Au was used as a cross-check since its mass is already known with an uncertainty...
The masses of 15 stable nuclides in the rare-earth region have been measured with the Penning-trap m...
The masses of 15 stable nuclides in the rare-earth region have been measured with the Penning-trap m...
Frequency ratio measurements with different combinations of the singly charged ions from 21, 22, 23N...
THe-Trap is a Penning-trap mass spectrometer at the Max-Planck-Institute for Nuclear Physics in Heid...
For a detailed study of the accuracy of the Penning trap mass spectrometer ISOLTRAP all expected sou...
ISOLTRAP is a Penning trap mass spectrometer installed at the on-line isotope separator ISOLDE/CERN....
Phase-imaging ion-cyclotron-resonance technique has been implemented at the Penning-trap mass spectr...
Phase-imaging ion-cyclotron-resonance technique has been implemented at the Penning-trap mass spectr...
The application of Penning traps for mass spectrometry has led to a major step in the mass precision...
Significant systematic errors in high-precision Penning trap mass spectrometry can result from elect...
The masses of 15 stable nuclides in the rare-earth region have been measured with the Penning-trap m...
The application of Penning traps for mass spectrometry has led to a major step in the mass precision...
The application of Penning traps for mass spectrometry has led to a major step in the mass precision...
Frequency ratio measurements with different combinations of the singly charged ions from 21, 22, 23N...
The nuclear masses reflect the sum of all interactions inside a nucleus. Their precise knowledge can...
The masses of 15 stable nuclides in the rare-earth region have been measured with the Penning-trap m...
The masses of 15 stable nuclides in the rare-earth region have been measured with the Penning-trap m...
Frequency ratio measurements with different combinations of the singly charged ions from 21, 22, 23N...
THe-Trap is a Penning-trap mass spectrometer at the Max-Planck-Institute for Nuclear Physics in Heid...
For a detailed study of the accuracy of the Penning trap mass spectrometer ISOLTRAP all expected sou...
ISOLTRAP is a Penning trap mass spectrometer installed at the on-line isotope separator ISOLDE/CERN....
Phase-imaging ion-cyclotron-resonance technique has been implemented at the Penning-trap mass spectr...
Phase-imaging ion-cyclotron-resonance technique has been implemented at the Penning-trap mass spectr...
The application of Penning traps for mass spectrometry has led to a major step in the mass precision...
Significant systematic errors in high-precision Penning trap mass spectrometry can result from elect...
The masses of 15 stable nuclides in the rare-earth region have been measured with the Penning-trap m...
The application of Penning traps for mass spectrometry has led to a major step in the mass precision...
The application of Penning traps for mass spectrometry has led to a major step in the mass precision...
Frequency ratio measurements with different combinations of the singly charged ions from 21, 22, 23N...
The nuclear masses reflect the sum of all interactions inside a nucleus. Their precise knowledge can...
The masses of 15 stable nuclides in the rare-earth region have been measured with the Penning-trap m...
The masses of 15 stable nuclides in the rare-earth region have been measured with the Penning-trap m...
Frequency ratio measurements with different combinations of the singly charged ions from 21, 22, 23N...