Using Goldmann applanation tonometry (GAT) as a gold standard, we evaluated the accuracy of Nidek NT-4000 pneumotonometry (NPT) in adults without corneal disease. Bland and Altman analysis of serial intra-ocular pressures (IOPs) measured with NPT and GAT in 10 healthy subjects revealed that the repeatability coefficients for the two methods were similar. NPT, GAT and ultrasonic pachymetry were then performed in 100 patients. Bland and Altman analysis showed that NPT yielded significantly higher readings than GAT [mean biases for right and left eye measurements were 1.37 mmHg (95% limits of agreement: -3.02-5.76) and 1.17 mmHg (95% limits of agreement: -2.76-5.11) respectively] and was more affected by corneal thickness variations. For detec...
AbstractTranspalpebral tonometer, a new generation intraocular pressure (IOP) measuring instrument i...
Purpose: To evaluate the agreement between the reading values of the Goldmann applanation tonometer ...
Background: To evaluate the agreement of intraocular pressure (IOP) and central corneal thickness (C...
Using Goldmann applanation tonometry (GAT) as a gold standard, we evaluated the accuracy of Nidek NT...
Purpose. This study compared a new noncontact tonometer (NCT), the Nidek NT-4.000 (Nidek Co., Ltd., ...
Sarwat Salim, Daniel J Linn, James R Echols II, Peter A NetlandHamilton Eye Institute, University of...
Ihsan Yilmaz, Cigdem Altan, Ebru Demet Aygit, Cengiz Alagoz, Okkes Baz, Sibel Ahmet, Semih Urvasizog...
OBJECTIVE: To assess the agreement of tonometers available for clinical practice with the Goldmann a...
Abstract Background To compare the measurement of intraocular pressure (IOP) among the three differe...
Purpose : To compare the level of accuracy of intraocular pressure (IOP) measurements attained by no...
Introduction: Goldmann applanation tonometer (GAT) is the gold standard for Intraocular Pressure (IO...
Background: The purpose of this study was to assess the repeatability and reproducibility of the PT1...
AIM: To compare intraocular pressure (IOP) values obtained from two different puff modes of Canon TX...
Purpose: To determine the intra- and inter-examiner repeatability of a new eyelid tonometer, and its...
AbstractIntroductionGoldmann applanation tonometer (GAT) is the gold standard for Intraocular Pressu...
AbstractTranspalpebral tonometer, a new generation intraocular pressure (IOP) measuring instrument i...
Purpose: To evaluate the agreement between the reading values of the Goldmann applanation tonometer ...
Background: To evaluate the agreement of intraocular pressure (IOP) and central corneal thickness (C...
Using Goldmann applanation tonometry (GAT) as a gold standard, we evaluated the accuracy of Nidek NT...
Purpose. This study compared a new noncontact tonometer (NCT), the Nidek NT-4.000 (Nidek Co., Ltd., ...
Sarwat Salim, Daniel J Linn, James R Echols II, Peter A NetlandHamilton Eye Institute, University of...
Ihsan Yilmaz, Cigdem Altan, Ebru Demet Aygit, Cengiz Alagoz, Okkes Baz, Sibel Ahmet, Semih Urvasizog...
OBJECTIVE: To assess the agreement of tonometers available for clinical practice with the Goldmann a...
Abstract Background To compare the measurement of intraocular pressure (IOP) among the three differe...
Purpose : To compare the level of accuracy of intraocular pressure (IOP) measurements attained by no...
Introduction: Goldmann applanation tonometer (GAT) is the gold standard for Intraocular Pressure (IO...
Background: The purpose of this study was to assess the repeatability and reproducibility of the PT1...
AIM: To compare intraocular pressure (IOP) values obtained from two different puff modes of Canon TX...
Purpose: To determine the intra- and inter-examiner repeatability of a new eyelid tonometer, and its...
AbstractIntroductionGoldmann applanation tonometer (GAT) is the gold standard for Intraocular Pressu...
AbstractTranspalpebral tonometer, a new generation intraocular pressure (IOP) measuring instrument i...
Purpose: To evaluate the agreement between the reading values of the Goldmann applanation tonometer ...
Background: To evaluate the agreement of intraocular pressure (IOP) and central corneal thickness (C...