A comparison of stability of chemical analytes in plasma from the BD Vacutainer (R) Barricor (TM) tube with mechanical separator versus tubes containing gel separator

  • Gawria, Ghassaan
  • Tillmar, Linda
  • Landberg, Eva
Publication date
January 2020
Publisher
Wiley

Abstract

Background: There is a need of prolonged stability of certain chemical analytes in lithium heparin tubes with separators. A new tube with a mechanical separator has recently been launched (Barricor (TM)), which according to the manufacturer may have these benefits. The aim of this study was to evaluate stability performance of this tube in comparison with plasma gel tubes under clinically realistic circumstances. Methods: Blood was collected in tubes containing lithium heparin with different separators; gel separator (Vacutainer (R) PST (TM), Becton Dickinson and Vacuette (R), Greiner bio-one) and mechanical separator (Vacutainer (R) Barricor (TM), Becton Dickinson). All tubes had an aspiration volume of 3 mL and were centrifuged at similar...