Abstract Background To investigate geometric and dosimetric accuracy of frame-less image-guided radiosurgery (IG-RS) for brain metastases. Methods and materials Single fraction IG-RS was practiced in 72 patients with 98 brain metastases. Patient positioning and immobilization used either double- (n = 71) or single-layer (n = 27) thermoplastic masks. Pre-treatment set-up errors (n = 98) were evaluated with cone-beam CT (CBCT) based image-guidance (IG) and were corrected in six degrees of freedom without an action level. CBCT imaging after treatment measured intra-fractional errors (n = 64). Pre- and post-treatment errors were simulated in the treatment planning system and target coverage and dose conformity were evaluated. Three scenarios of...
Introduction Stereotactic radiosurgery (SRS) has been increasingly advocated for 1–3 small brain met...
Purpose/Objective(s): Stereotactic radiosurgery with image guidance is becoming widely adopted for t...
Background/Purpose: This study determines and compares the geometric setup errors between stereoscop...
Background: To investigate geometric and dosimetric accuracy of frame-less image-guided radiosurgery...
Abstract Purpose To investigate the geometric and dosimetric impacts of intra-fractional movement fo...
Abstract Purpose To assess the accuracy of patient repositioning and clinical outcomes of frameless ...
The accuracy of radiation delivery is increasingly important as radiotherapy technology continues to...
Hintergrund: Kranielle Stereotaxie ist ein wichtiges Therapieinstrument zur Behandlung kranieller ne...
Introduction: Inherent to SRS and the small treatment fields used are the tight margins and resultin...
Purpose: Mask-immobilized stereotactic radiosurgery (SRS) using a gating window is an emerging techn...
Stereotactic radiotherapy (SRT) requires tight margins around the tumor, thus producing a steep dose...
The aim of this study was to evaluate the impact of actual rotational setup errors on dose distribut...
AbstractAimTo analyze intrafraction movement in patients undergoing frameless robotic radiosurgery a...
<p><bold>Purpose:</bold> Stereotactic Radiosurgery (SRS) indicates a single fraction, large prescrip...
To assess interfraction translational and rotational setup errors, in patients treated with image-...
Introduction Stereotactic radiosurgery (SRS) has been increasingly advocated for 1–3 small brain met...
Purpose/Objective(s): Stereotactic radiosurgery with image guidance is becoming widely adopted for t...
Background/Purpose: This study determines and compares the geometric setup errors between stereoscop...
Background: To investigate geometric and dosimetric accuracy of frame-less image-guided radiosurgery...
Abstract Purpose To investigate the geometric and dosimetric impacts of intra-fractional movement fo...
Abstract Purpose To assess the accuracy of patient repositioning and clinical outcomes of frameless ...
The accuracy of radiation delivery is increasingly important as radiotherapy technology continues to...
Hintergrund: Kranielle Stereotaxie ist ein wichtiges Therapieinstrument zur Behandlung kranieller ne...
Introduction: Inherent to SRS and the small treatment fields used are the tight margins and resultin...
Purpose: Mask-immobilized stereotactic radiosurgery (SRS) using a gating window is an emerging techn...
Stereotactic radiotherapy (SRT) requires tight margins around the tumor, thus producing a steep dose...
The aim of this study was to evaluate the impact of actual rotational setup errors on dose distribut...
AbstractAimTo analyze intrafraction movement in patients undergoing frameless robotic radiosurgery a...
<p><bold>Purpose:</bold> Stereotactic Radiosurgery (SRS) indicates a single fraction, large prescrip...
To assess interfraction translational and rotational setup errors, in patients treated with image-...
Introduction Stereotactic radiosurgery (SRS) has been increasingly advocated for 1–3 small brain met...
Purpose/Objective(s): Stereotactic radiosurgery with image guidance is becoming widely adopted for t...
Background/Purpose: This study determines and compares the geometric setup errors between stereoscop...