International audienceReference values of hemodynamic parameters for the assessment of prosthetic heart valves are necessary, and ideally need to be provided by entities independent of the valve manufacturers. Thus, the study aim was to provide, in vitro, normal reference values of the effective orifice area (EOA) for different models and sizes of mitral prosthetic valve, and to assess the determinants of EOA and mean transvalvular pressure gradient (mTPG)
The inherent limitations of current heart valve substitutes create the premise for the Tissue Engine...
We studied and compared functional parameters in 314 valvular prostheses. The following parameters h...
Bogunovic N, Horstkotte D, Faber L, et al. Effective Valve Opening Area in the Detection of Dysfunct...
International audienceReference values of hemodynamic parameters for the assessment of prosthetic he...
Objectives: Reference values of hemodynamic parameters for the assessment of prosthetic heart valves...
The relation between the geometrical orifice area (GOA) and the effective orifice area (EOA) of mitr...
Bogunovic N, Horstkotte D, Faber L, Bogunovic L, van Buuren F. Echocardiographically derived effecti...
International audienceEffective orifice area (EOA) and mean pressure gradientare routine echocardiog...
The Gorlin equation for the hemodynamic assessment of valve area is commonly used in cardiac cathete...
Objective This study compared the continuity equation-based effective orifice area (EOA) of prosthet...
AbstractObjectives: We sought to evaluate the Doppler assessment of effective orifice area in aortic...
The Gorlin equation for the hemodynamic assessment of valve area is commonly used in cardiac cathete...
Doppler echocardiographic characteristics of normally functioning Allcarbon prostheses were studied ...
AbstractTo assess the effect of fluid flow on orifice area and to test the Gorlin formula, six Carpe...
The \u2018in vitro\u2019 assessement of hydrodynamic performance of cardiac valve prostheses is regu...
The inherent limitations of current heart valve substitutes create the premise for the Tissue Engine...
We studied and compared functional parameters in 314 valvular prostheses. The following parameters h...
Bogunovic N, Horstkotte D, Faber L, et al. Effective Valve Opening Area in the Detection of Dysfunct...
International audienceReference values of hemodynamic parameters for the assessment of prosthetic he...
Objectives: Reference values of hemodynamic parameters for the assessment of prosthetic heart valves...
The relation between the geometrical orifice area (GOA) and the effective orifice area (EOA) of mitr...
Bogunovic N, Horstkotte D, Faber L, Bogunovic L, van Buuren F. Echocardiographically derived effecti...
International audienceEffective orifice area (EOA) and mean pressure gradientare routine echocardiog...
The Gorlin equation for the hemodynamic assessment of valve area is commonly used in cardiac cathete...
Objective This study compared the continuity equation-based effective orifice area (EOA) of prosthet...
AbstractObjectives: We sought to evaluate the Doppler assessment of effective orifice area in aortic...
The Gorlin equation for the hemodynamic assessment of valve area is commonly used in cardiac cathete...
Doppler echocardiographic characteristics of normally functioning Allcarbon prostheses were studied ...
AbstractTo assess the effect of fluid flow on orifice area and to test the Gorlin formula, six Carpe...
The \u2018in vitro\u2019 assessement of hydrodynamic performance of cardiac valve prostheses is regu...
The inherent limitations of current heart valve substitutes create the premise for the Tissue Engine...
We studied and compared functional parameters in 314 valvular prostheses. The following parameters h...
Bogunovic N, Horstkotte D, Faber L, et al. Effective Valve Opening Area in the Detection of Dysfunct...