Acute and long-term pacing thresholds were measured prospectively in 74 patients with a unipolar/bipolar mul-tiprogrammable pacemaker. At implantation, mean current threshold was 0.48 ± 0.16 mA with unipolar mode and 0.55 ± 0.16 mA bipolar mode (p < 0.01). R wave amplitude at implantation was 7.78 ± 2.4 mV with unipolar and 7.67 ± 2.1 mV in bipolar mode (p = NS). During long-term follow-up (mean 9.3 months; range 3 to 24), no clinically significant differences in pacing or sensing thresholds were observed between bipolar and unipolar configurations.Lead configuration was changed 23 times in 11 patients. Symptomatic myopotential inhibition was corrected in two patients by reprogramming to the bipolar pacing mode. High thresholds and loss of ...
Thisstudy aims to compute 50 Hz electric field interferences on pacemakers for diverse lead configur...
This study examines the changes in pacing threshold and R- or P-wave amplitude during the first 30 m...
The continued efficacy of dual chamber pacing is predicated on the stability of both atrial and vent...
Acute and long-term pacing thresholds were measured prospectively in 74 patients with a unipolar/bip...
One hundred and seventy-three patients, mean age 74 years permanently paced with 123 atrial (53 unip...
Single lead DDD pacing using unipolar or bipolar stimulation is limited by high atrial threshold. Ov...
AIMS: The position of the pacemaker (PM) system (right or left-sided, pectoral, or abdominal), th...
PubMedID: 12494622The Autocapture function controls and optimizes the amplitude of the pacing pulse ...
The effects of current strength (threshold to 20 mA) and pacing polarity (bipolar versus unipolar) o...
Background: Active-fixation screw-in leads are associated with initially high thresholds. However, p...
In a group of 20 patients implanted with a single-lead VDD pacing system, a wide interindividual var...
\(\bf Background:\) Bipolar active fixation (BipolarAFL) and quadripolar passive fixation left-ventr...
A new pacing mode which significantly reduced atrial pacing threshold, the 'overlapping biphasic imp...
Background: The effect of left ventricular (LV) systolic function on the long-term left ventricular ...
Peer Reviewedhttp://deepblue.lib.umich.edu/bitstream/2027.42/73370/1/j.1540-8159.1991.tb04113.x.pd
Thisstudy aims to compute 50 Hz electric field interferences on pacemakers for diverse lead configur...
This study examines the changes in pacing threshold and R- or P-wave amplitude during the first 30 m...
The continued efficacy of dual chamber pacing is predicated on the stability of both atrial and vent...
Acute and long-term pacing thresholds were measured prospectively in 74 patients with a unipolar/bip...
One hundred and seventy-three patients, mean age 74 years permanently paced with 123 atrial (53 unip...
Single lead DDD pacing using unipolar or bipolar stimulation is limited by high atrial threshold. Ov...
AIMS: The position of the pacemaker (PM) system (right or left-sided, pectoral, or abdominal), th...
PubMedID: 12494622The Autocapture function controls and optimizes the amplitude of the pacing pulse ...
The effects of current strength (threshold to 20 mA) and pacing polarity (bipolar versus unipolar) o...
Background: Active-fixation screw-in leads are associated with initially high thresholds. However, p...
In a group of 20 patients implanted with a single-lead VDD pacing system, a wide interindividual var...
\(\bf Background:\) Bipolar active fixation (BipolarAFL) and quadripolar passive fixation left-ventr...
A new pacing mode which significantly reduced atrial pacing threshold, the 'overlapping biphasic imp...
Background: The effect of left ventricular (LV) systolic function on the long-term left ventricular ...
Peer Reviewedhttp://deepblue.lib.umich.edu/bitstream/2027.42/73370/1/j.1540-8159.1991.tb04113.x.pd
Thisstudy aims to compute 50 Hz electric field interferences on pacemakers for diverse lead configur...
This study examines the changes in pacing threshold and R- or P-wave amplitude during the first 30 m...
The continued efficacy of dual chamber pacing is predicated on the stability of both atrial and vent...