Purpose: The purpose of this study is to compare computed tomography (CT) radiation dose measurement methods proposed by TG111, International Electrotechnical Commission (IEC), and a direct dose profile integral (DPI) measurement method. Methods: Pencil and Farmer ion chambers are used for integrating dose profiles at different beam widths in a 60 cm long body phantom. Resulting DPI is used to calculate CT dose index (CTDI) at each beam width. Measurements are also done for a pencil chamber inserted into a 15 cm body phantom at the reference beam width. The reference measurement is scaled with pencil chamber measurements in air at different beam widths, according to the IEC approach. Finally, point dose measurements are done with a Farmer c...
Background: Medical X-rays are the largest man-made source of public exposure to ionizing radiation...
Current estimation of radiation dose from computed tomography (CT) scans on patients has relied on t...
Includes bibliographical references (pages 72-75).The simplest and most standardized metric for repo...
Purpose: The purpose of this study is to compare computed tomography (CT) radiation dose measurement...
ABSTRACT Background: Nowadays, the use of computed tomography (CT) as a diagnostic tool has been c...
Background: Internationally, the typical allowed difference between the measured radiation dose and ...
This paper reports the performance and dose assessment of the X-ray generator and CT dosimetry of 64...
The most common parameter used to estimate and minimise patient dose in computed tomography (CT) is ...
Objective: To survey computed tomography (CT) radiation dose associated with non-contrast spiral Mul...
[[abstract]]©2003 AIP - Comprehensive analyses and measurements of computed tomography (CT) single-s...
The method of CT dosimetry based on the CTDI is a common measurement method by using a single, axial...
Purpose: The weighted CT dose index (CTDIw) has been extended for a nominal total collimation width ...
Dose and image quality assessment in computed tomography (CT) are almost affected by the vast variet...
Introduction: Computed tomography (CT) has numerous applications in clinical procedures but its main...
[[abstract]]©2007 AIP - The IAEA and the ICRP recommended dose guidance levels for the most frequent...
Background: Medical X-rays are the largest man-made source of public exposure to ionizing radiation...
Current estimation of radiation dose from computed tomography (CT) scans on patients has relied on t...
Includes bibliographical references (pages 72-75).The simplest and most standardized metric for repo...
Purpose: The purpose of this study is to compare computed tomography (CT) radiation dose measurement...
ABSTRACT Background: Nowadays, the use of computed tomography (CT) as a diagnostic tool has been c...
Background: Internationally, the typical allowed difference between the measured radiation dose and ...
This paper reports the performance and dose assessment of the X-ray generator and CT dosimetry of 64...
The most common parameter used to estimate and minimise patient dose in computed tomography (CT) is ...
Objective: To survey computed tomography (CT) radiation dose associated with non-contrast spiral Mul...
[[abstract]]©2003 AIP - Comprehensive analyses and measurements of computed tomography (CT) single-s...
The method of CT dosimetry based on the CTDI is a common measurement method by using a single, axial...
Purpose: The weighted CT dose index (CTDIw) has been extended for a nominal total collimation width ...
Dose and image quality assessment in computed tomography (CT) are almost affected by the vast variet...
Introduction: Computed tomography (CT) has numerous applications in clinical procedures but its main...
[[abstract]]©2007 AIP - The IAEA and the ICRP recommended dose guidance levels for the most frequent...
Background: Medical X-rays are the largest man-made source of public exposure to ionizing radiation...
Current estimation of radiation dose from computed tomography (CT) scans on patients has relied on t...
Includes bibliographical references (pages 72-75).The simplest and most standardized metric for repo...