Silicon-diode-based detectors are commonly used for the dosimetry of small radiotherapy beams due to their relatively small volumes and high sensitivity to ionizing radiation. Nevertheless, silicon-diode-based detectors tend to over-respond in small fields because of their high density relative to water. For that reason, detector-specific beam correction factors ([Formula: see text]) have been recommended not only to correct the total scatter factors but also to correct the tissue maximum and off-axis ratios. However, the application of [Formula: see text] to in-depth and off-axis locations has not been studied. The goal of this work is to address the impact of the correction factors on the calculated dose distribution in static non-convent...
Accurate dosimetry is a fundamental requirement for the safe and efficient use of radiation in medic...
A coherent system for the use of scatter correction factors, determined at 10 cm depth, is described...
A coherent system for the use of scatter correction factors, determined at 10 cm depth, is described...
<div><p>Silicon-diode-based detectors are commonly used for the dosimetry of small radiotherapy beam...
<p>Correction factors to TMR for circular collimators of a) 4.0 mm, b) 10.0 mm, and c) 20.0 mm at t...
Purpose: The aim of the present study is to provide a comprehensive set of detector specific correct...
For each radiation treatment machine, the dose per monitor unit must be known at a reference point i...
Stereotactic radiosurgery (SRS) is a technique where an external photon beam is used to deliver a h...
Radiosurgery is characterized by high radiation doses, delivered via small diameter radiation beams ...
International audiencePurpose The use of small photon fields is now an established practice in stere...
International audiencePurpose The use of small photon fields is now an established practice in stere...
Accurate dosimetry is a fundamental requirement for the safe and efficient use of radiation in medic...
International audiencePurpose The use of small photon fields is now an established practice in stere...
International audiencePurpose The use of small photon fields is now an established practice in stere...
International audiencePurpose The use of small photon fields is now an established practice in stere...
Accurate dosimetry is a fundamental requirement for the safe and efficient use of radiation in medic...
A coherent system for the use of scatter correction factors, determined at 10 cm depth, is described...
A coherent system for the use of scatter correction factors, determined at 10 cm depth, is described...
<div><p>Silicon-diode-based detectors are commonly used for the dosimetry of small radiotherapy beam...
<p>Correction factors to TMR for circular collimators of a) 4.0 mm, b) 10.0 mm, and c) 20.0 mm at t...
Purpose: The aim of the present study is to provide a comprehensive set of detector specific correct...
For each radiation treatment machine, the dose per monitor unit must be known at a reference point i...
Stereotactic radiosurgery (SRS) is a technique where an external photon beam is used to deliver a h...
Radiosurgery is characterized by high radiation doses, delivered via small diameter radiation beams ...
International audiencePurpose The use of small photon fields is now an established practice in stere...
International audiencePurpose The use of small photon fields is now an established practice in stere...
Accurate dosimetry is a fundamental requirement for the safe and efficient use of radiation in medic...
International audiencePurpose The use of small photon fields is now an established practice in stere...
International audiencePurpose The use of small photon fields is now an established practice in stere...
International audiencePurpose The use of small photon fields is now an established practice in stere...
Accurate dosimetry is a fundamental requirement for the safe and efficient use of radiation in medic...
A coherent system for the use of scatter correction factors, determined at 10 cm depth, is described...
A coherent system for the use of scatter correction factors, determined at 10 cm depth, is described...