A collimated source of 241Am was scanned over the endcap of a 21 year old coaxial HPGe-detector that had spent about 75% of its life at room temperature (and the remaining time at 77 K). The detector response was recorded and used as a measure of the relative thickness of the top deadlayer. This thickness was not homogeneous and was thicker near to the outer surface of the crystal compared to the centre, which could be a result of increased diffusion of Li atoms during times the detector was kept at room temperature. The results were compared with two newer HPGe-detectors that proved to have homogeneous top deadlayers.JRC.G.2-Standards for Nuclear Safety, Security and Safeguard
The performance of a Compton-suppression spectrometer is investigated for two different Ge crystals ...
In this project, the homogeneity of the detection efficiency of two silicon detectors were examined ...
There is some possibility that many granular regions which have a Cl density approximately one order...
The dead-layer uniformity of the top surface of two high purity germanium detectors has been studied...
High-purity germanium (HPGe) detectors are commonly used for high-resolution gamma spectrometry. A H...
In this study, Monte Carlo simulations were used to calculate the full-energy peak efficiency of a p...
One important factor in using an High Purity Germanium (HPGe) detector is its efficiency that highly...
This Dissertation is brought to you for free and open access by the Graduate School at Trace: Tennes...
The effects of different passivation methods applied to the same planar high-purity germanium gamma ...
The measurement capacity of the HADES underground laboratory was recently increased with a fourth de...
AbstractWe report on the efficiency calibration of a HPGe x-ray detector using radioactive sources a...
A decommissioned p-type, co-axial HPGe detector has been recently brought back to operational mode i...
Computed Tomography (CT) high resolution imaging have been used to investigate if there is a signifi...
Since a few years there are well-type HPGe-detectors with a small, point-like, anode contacts availa...
Six HPGe detectors with various crystal diameters were investigated for data related to the aspect r...
The performance of a Compton-suppression spectrometer is investigated for two different Ge crystals ...
In this project, the homogeneity of the detection efficiency of two silicon detectors were examined ...
There is some possibility that many granular regions which have a Cl density approximately one order...
The dead-layer uniformity of the top surface of two high purity germanium detectors has been studied...
High-purity germanium (HPGe) detectors are commonly used for high-resolution gamma spectrometry. A H...
In this study, Monte Carlo simulations were used to calculate the full-energy peak efficiency of a p...
One important factor in using an High Purity Germanium (HPGe) detector is its efficiency that highly...
This Dissertation is brought to you for free and open access by the Graduate School at Trace: Tennes...
The effects of different passivation methods applied to the same planar high-purity germanium gamma ...
The measurement capacity of the HADES underground laboratory was recently increased with a fourth de...
AbstractWe report on the efficiency calibration of a HPGe x-ray detector using radioactive sources a...
A decommissioned p-type, co-axial HPGe detector has been recently brought back to operational mode i...
Computed Tomography (CT) high resolution imaging have been used to investigate if there is a signifi...
Since a few years there are well-type HPGe-detectors with a small, point-like, anode contacts availa...
Six HPGe detectors with various crystal diameters were investigated for data related to the aspect r...
The performance of a Compton-suppression spectrometer is investigated for two different Ge crystals ...
In this project, the homogeneity of the detection efficiency of two silicon detectors were examined ...
There is some possibility that many granular regions which have a Cl density approximately one order...