Within the ORAMED project (Optimization of Radiation Protection of Medical Staff) a coordinated measurement program for occupationally exposed medical staff was performed in different hospitals in Europe (www.oramed-fp7.eu). The main objective was to obtain a set of standardized data on extremity and eye lens doses for staff involved in interventional radiology and cardiology and to optimize radiation protection. Special attention was given to the measurement of the doses to the eye lenses. In this paper an overview will be given of the measured eye lens doses and the main influence factors for these doses. The measured eye lens doses are extrapolated to annual doses. The extrapolations showed that monitoringof the eye lens should be perfor...
Objective: Early 2018, the new eye lens dose limit of 20 mSv per year for occupational exposure to i...
Introduction: Medical professionals performing fluoroscopy guided interventional procedures in cardi...
The purpose of this paper is to test the appropriateness of OSL and electronic dosemeters to estimat...
The main objective of WP1 of the ORAMED (Optimization of RAdiation protection for MEDical staff) pro...
The main aim of the Work Package 1 (WP1) of the ORAMED project, Collaborative Project (2008-2011), s...
Workers involved in interventional cardiology procedures receive high eye lens dose if protection is...
Medical personnel performing interventional procedures in cardiology and radiology is considered to ...
Objective: This study measured the radiation exposure of the eye lens of medical personnel performin...
Interventional radiology and cardiology are areas with high potential for risk to eye lens. Accurate...
Occupational exposure from interventional x-ray procedures is one of the areas in which increased ey...
A thorough literature review about the current situation on the implementation of eye lens monitorin...
A thorough literature review about the current situation on the implementation of eye lens monitorin...
Occupational exposure from interventional x-ray procedures is one of the areas in which increased ey...
Occupational exposure from interventional x-ray procedures is one of the areas in which increased ey...
The equivalent dose limit for the eye lens for occupational exposure recommended by the ICRP has bee...
Objective: Early 2018, the new eye lens dose limit of 20 mSv per year for occupational exposure to i...
Introduction: Medical professionals performing fluoroscopy guided interventional procedures in cardi...
The purpose of this paper is to test the appropriateness of OSL and electronic dosemeters to estimat...
The main objective of WP1 of the ORAMED (Optimization of RAdiation protection for MEDical staff) pro...
The main aim of the Work Package 1 (WP1) of the ORAMED project, Collaborative Project (2008-2011), s...
Workers involved in interventional cardiology procedures receive high eye lens dose if protection is...
Medical personnel performing interventional procedures in cardiology and radiology is considered to ...
Objective: This study measured the radiation exposure of the eye lens of medical personnel performin...
Interventional radiology and cardiology are areas with high potential for risk to eye lens. Accurate...
Occupational exposure from interventional x-ray procedures is one of the areas in which increased ey...
A thorough literature review about the current situation on the implementation of eye lens monitorin...
A thorough literature review about the current situation on the implementation of eye lens monitorin...
Occupational exposure from interventional x-ray procedures is one of the areas in which increased ey...
Occupational exposure from interventional x-ray procedures is one of the areas in which increased ey...
The equivalent dose limit for the eye lens for occupational exposure recommended by the ICRP has bee...
Objective: Early 2018, the new eye lens dose limit of 20 mSv per year for occupational exposure to i...
Introduction: Medical professionals performing fluoroscopy guided interventional procedures in cardi...
The purpose of this paper is to test the appropriateness of OSL and electronic dosemeters to estimat...