Sir, The body fluid (BF) module on the XN-1000 (Sysmex Corporation, Kobe, Japan) represents a convenient platform for accurate and fast quantification of white blood cells (WBC) by flow cytometry in multiple types of BF. Despite their well-known advantages in comparison with light microscopy, analyzers with BF mode can be affected by different types of interference leading to spurious results and potential clinical misinterpretations [...
markdownabstractCellular analysis of body fluids, including cerebrospinal fluid and serous fluids, p...
Background: Caution is of paramount importance in the interpretation of flow cytometric white blood ...
Background: We evaluated the new body fluid module on Sysmex UF1000i (UF1000i-BF) for analysis of wh...
INTRODUCTION: Cellular analysis in cerebrospinal fluid (CSF) provides important diagnostic informat...
<div><p>The XN series automated hematology analyzer has been equipped with a body fluid (BF) mode to...
The XN series automated hematology analyzer has been equipped with a body fluid (BF) mode to count a...
The body fluid mode of the Sysmex XN-2000 hematology analyzer differentiates cells into mononuclear ...
The presence of high fluorescent cells high fluorescent cells (on the body fluid analyzer) (HF-BF) o...
INTRODUCTION. Cellular analysis in body fluids (BFs) provides important diagnostic information in va...
Background: This study was planned to evaluate the analytical performance of the novel and fully aut...
Morphological microscopic examinations of nucleated cells in body fluid (BF) samples are performed t...
Introduction: In 2015, Sysmex launched a new series of hematology analyzers (XN-L Series) designed t...
In many inflammatory diseases, the cellular components in body fluids [cerebrospinal fluid (CSF), se...
New developments in flow cytometry are now being applied in toxicology studies. There are several re...
Simulated candidemia was produced with 20 Candida isolates at three inoculum sizes (100, 10, and 1 C...
markdownabstractCellular analysis of body fluids, including cerebrospinal fluid and serous fluids, p...
Background: Caution is of paramount importance in the interpretation of flow cytometric white blood ...
Background: We evaluated the new body fluid module on Sysmex UF1000i (UF1000i-BF) for analysis of wh...
INTRODUCTION: Cellular analysis in cerebrospinal fluid (CSF) provides important diagnostic informat...
<div><p>The XN series automated hematology analyzer has been equipped with a body fluid (BF) mode to...
The XN series automated hematology analyzer has been equipped with a body fluid (BF) mode to count a...
The body fluid mode of the Sysmex XN-2000 hematology analyzer differentiates cells into mononuclear ...
The presence of high fluorescent cells high fluorescent cells (on the body fluid analyzer) (HF-BF) o...
INTRODUCTION. Cellular analysis in body fluids (BFs) provides important diagnostic information in va...
Background: This study was planned to evaluate the analytical performance of the novel and fully aut...
Morphological microscopic examinations of nucleated cells in body fluid (BF) samples are performed t...
Introduction: In 2015, Sysmex launched a new series of hematology analyzers (XN-L Series) designed t...
In many inflammatory diseases, the cellular components in body fluids [cerebrospinal fluid (CSF), se...
New developments in flow cytometry are now being applied in toxicology studies. There are several re...
Simulated candidemia was produced with 20 Candida isolates at three inoculum sizes (100, 10, and 1 C...
markdownabstractCellular analysis of body fluids, including cerebrospinal fluid and serous fluids, p...
Background: Caution is of paramount importance in the interpretation of flow cytometric white blood ...
Background: We evaluated the new body fluid module on Sysmex UF1000i (UF1000i-BF) for analysis of wh...