The aim of this study was to determine the accuracy of the WatchBP Office ABI monitor for office blood pressure measurement over a wide range of arm circumferences using the ANSI/AAMI/ISO 81060-2:2013 protocol. The device accuracy was tested in 88 participants whose mean\ub1SD age was 54.5\ub117.6 years, whose arm circumference was 30.6\ub18.3\u2009cm (range: 15-46\u2009cm), and whose entry blood pressure (BP) was 138.3\ub123.4\u2009mmHg for systolic and 83.7\ub114.6\u2009mmHg for diastolic BP. Four cuffs (small, standard, large, and extra-large) suitable for arm circumferences ranging from 14.0 to 52.0\u2009cm were used. The mean device-observer difference in the 264 separate BP data pairs was 0.7\ub13.8\u2009mmHg for systolic BP and was 0...
Measurement of blood pressure (BP) using the auscultatory method must follow specific rules and cond...
Background: Independent testing of home blood pressure (BP) measurement (HBPM) devices is often lack...
OBJECTIVE: Automated cuff blood pressure (BP) devices are widely used for ambulatory, home, and offi...
OBJECTIVE: To determine the accuracy of the WatchBP Office ABI monitor for blood pressure measure...
The aim of this study was to determine the accuracy of the A&D UM-201 device coupled to several cuff...
To determine the accuracy of the large-extra large-sized (L-XL) cuff (32-52 cm) coupled to a Microli...
The aim of this study was to define the accuracy of UM-211, an automated oscillometric device for of...
Despite the increasing popularity of blood pressure (BP) wrist monitors for self-BP measurement at h...
Hypertension is the most significant modifiable risk factor for cardiovascular disease and contributes...
Objectives: Handheld medical devices and smartwatches that measure BP without a cuff have recently b...
Background Despite the increasing popularity of wrist-cuff blood pressure (BP) devices, their accura...
OBJECTIVE: To determine the accuracy of the Omron RX-M, a device measuring blood pressure oscillomet...
OBJECTIVE: The objective of this study was to determine the accuracy of the A&D BP UB-542 wrist ...
Objective: Extensive marketing of devices for self-measurement of blood pressure has created a need ...
Although they have been marketed widely, few data about the diagnostic accuracy of blood pressure mo...
Measurement of blood pressure (BP) using the auscultatory method must follow specific rules and cond...
Background: Independent testing of home blood pressure (BP) measurement (HBPM) devices is often lack...
OBJECTIVE: Automated cuff blood pressure (BP) devices are widely used for ambulatory, home, and offi...
OBJECTIVE: To determine the accuracy of the WatchBP Office ABI monitor for blood pressure measure...
The aim of this study was to determine the accuracy of the A&D UM-201 device coupled to several cuff...
To determine the accuracy of the large-extra large-sized (L-XL) cuff (32-52 cm) coupled to a Microli...
The aim of this study was to define the accuracy of UM-211, an automated oscillometric device for of...
Despite the increasing popularity of blood pressure (BP) wrist monitors for self-BP measurement at h...
Hypertension is the most significant modifiable risk factor for cardiovascular disease and contributes...
Objectives: Handheld medical devices and smartwatches that measure BP without a cuff have recently b...
Background Despite the increasing popularity of wrist-cuff blood pressure (BP) devices, their accura...
OBJECTIVE: To determine the accuracy of the Omron RX-M, a device measuring blood pressure oscillomet...
OBJECTIVE: The objective of this study was to determine the accuracy of the A&D BP UB-542 wrist ...
Objective: Extensive marketing of devices for self-measurement of blood pressure has created a need ...
Although they have been marketed widely, few data about the diagnostic accuracy of blood pressure mo...
Measurement of blood pressure (BP) using the auscultatory method must follow specific rules and cond...
Background: Independent testing of home blood pressure (BP) measurement (HBPM) devices is often lack...
OBJECTIVE: Automated cuff blood pressure (BP) devices are widely used for ambulatory, home, and offi...