Purpose. To assess the repeatability and agreement of parameters obtained with two biometers and to compare the predictability. Methods. Biometry was performed on 101 eyes with cataract using the IOLMaster 700 and the Galilei G6. Three measurements were obtained per eye with each device, and repeatability was evaluated. The axial length (AL), anterior chamber depth (ACD), keratometry (K), white-to-white (WTW) corneal diameter, central corneal thickness (CCT), and lens thickness (LT) were measured and postoperative predictability was compared. Results. Measurements could not be obtained with the IOLMaster 700 in one eye and in seven eyes with the Galilei G6 due to dense cataract. Both the IOLMaster 700 and Galilei G6 showed good repeatabilit...
Purpose. To comprehensively assess the reliability of a new optical biometer (IOLMaster 700), based ...
Purpose: To compare the clinical measurements with the new IOLMaster-700 from those obtained with th...
Purpose Accurate biometry is an obligatory preoperative measurement for refractive surgery as wel...
Purpose: To compare the repeatability of ocular biometry measured with the LENSTAR LS 900, IOLMaster...
Purpose: To compare the repeatability of ocular biometry measured with the LENSTAR LS 900, IOLMaster...
Purpose: To compare the repeatability of ocular biometry measured with the LENSTAR LS 900, IOLMaster...
Purpose: To compare the repeatability of ocular biometry measured with the LENSTAR LS 900, IOLMaster...
Purpose To evaluate the repeatability and reproducibility of measurements obtained by a new optical ...
Purpose To compare measurements taken using a swept-source optical coherence tomography-based optica...
Purpose: To evaluate the agreement between two biometry devices, the Heidelberg Anterion and the Gal...
Samuel F Passi,* Atalie C Thompson,* Preeya K Gupta Department of Ophthalmology, Duke University Ey...
This research aimed to investigate the potential differences in the parameters, including axial leng...
PurposeTo evaluate the repeatability and reproducibility of a newer swept-source optical biometer an...
Purpose:To evaluate the repeatability and agreement of the anterior segment measurements obtained us...
PURPOSE: To evaluate the repeatability and reproducibility of measurements obtained by a new optical...
Purpose. To comprehensively assess the reliability of a new optical biometer (IOLMaster 700), based ...
Purpose: To compare the clinical measurements with the new IOLMaster-700 from those obtained with th...
Purpose Accurate biometry is an obligatory preoperative measurement for refractive surgery as wel...
Purpose: To compare the repeatability of ocular biometry measured with the LENSTAR LS 900, IOLMaster...
Purpose: To compare the repeatability of ocular biometry measured with the LENSTAR LS 900, IOLMaster...
Purpose: To compare the repeatability of ocular biometry measured with the LENSTAR LS 900, IOLMaster...
Purpose: To compare the repeatability of ocular biometry measured with the LENSTAR LS 900, IOLMaster...
Purpose To evaluate the repeatability and reproducibility of measurements obtained by a new optical ...
Purpose To compare measurements taken using a swept-source optical coherence tomography-based optica...
Purpose: To evaluate the agreement between two biometry devices, the Heidelberg Anterion and the Gal...
Samuel F Passi,* Atalie C Thompson,* Preeya K Gupta Department of Ophthalmology, Duke University Ey...
This research aimed to investigate the potential differences in the parameters, including axial leng...
PurposeTo evaluate the repeatability and reproducibility of a newer swept-source optical biometer an...
Purpose:To evaluate the repeatability and agreement of the anterior segment measurements obtained us...
PURPOSE: To evaluate the repeatability and reproducibility of measurements obtained by a new optical...
Purpose. To comprehensively assess the reliability of a new optical biometer (IOLMaster 700), based ...
Purpose: To compare the clinical measurements with the new IOLMaster-700 from those obtained with th...
Purpose Accurate biometry is an obligatory preoperative measurement for refractive surgery as wel...